If you would like to know more on the matter you can look at ANSI/NCSL Z540-1-1994. That is the spec that calibration labs have to follow. You can use a NIST STANDARD or transfer standard for your instrument  calibrations...either way there is a 4:1 test accuracy ratio simply meaning whatever instrument you are calibrating  has to be done by another instrument that is 4 times more accurate. And talk about compounding percentages that spec also  says that the collective uncertainty of the measuring standards shall not exceed 25% of acceptable tolerance for each characteristic. lol.